STXL-6303 Possible Sensor Damage

Discussion in 'STX and STXL Series Cameras' started by Sam Medina, Oct 29, 2025.

  1. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    I acquired a STXL-6303E and during the check out I noticed that the Bias image was not that quite right. I included both a Bias and Dark image (along with a screen shot)from the camera.

    I would like to see if it is possible to have it repaired if needed.

    Thanks,

    Sam
     

    Attached Files:

  2. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    8,969
    Location:
    Earth
    Sam, your FITS files are not loadable. Please try again with CCDOps or MaxIm DL Pro.
    Please explain your test setup.
     
  3. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    My setup was a simple bench test. The camera was connected directly to my PC via USB2 cable (that cable was replaced to ensure that was that the issue), I opened Maxim DL Pro connected to it and sensor temp was set to -20C.

    The same results were obtained with SB TheSkyX Pro.

    Sorry about the corrupted files, I ensure that both of these were opened with Maxim DL Pro and PixInSight.
     

    Attached Files:

  4. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    8,969
    Location:
    Earth
    Thanks for the files.
    "A simple bench test" doesn't tell me much.
    What's the serial number of the camera?
    Was the camera covered up (cap on) to prevent light from leaking in? In a dark place?
    Have you looked down the nose toward the sensor to see if the shutter is working? eg if you expose a light frame for 5s), it should close then open for 5s then close.
    The bias frame is hard to understand - it partly looks like a shutter problem.
    Is there a filter wheel attached?

    If you want to send it in, we'll need some of that information. Please fill in this form, and my colleague @Tim will contact you about sending it in to Ottawa.
    https://diffractionlimited.com/request-rma/
     
  5. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    10,801
    Looks like a massive light leak.

    Note that the shutter can't block out daylight by itself - the camera is so sensitive it will pick up the tiny amount of light that is bouncing around the shutter. Put a cap on the camera and try again.
     
  6. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    Gentlemen,

    The camera has a FW attached to it and I have ensured that it is securely attached. The following is the information requested.

    Camera - Model: STXL-6303, Serial#: X13090268
    CFW - Model: FW8-STXL, Serial#: 1306

    Software utilized for the test: Maxim DL Version 7.22
    Camera Temp during test -20C

    Shutter was operational (during the 5 sec light frame, fit included)

    2 Bias Fits attached (with cap on and without cap on the FW))
    2 Dark Fits attached (300 seconds with cap on and without cap on the FW)

    Thanks in advance,

    Sam
     

    Attached Files:

  7. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    10,801
    Is it possible that your CCD chamber is saturated with moisture? That can cause all sorts of problems.

    I'd recommend recharging the desiccant, waiting 24 hours, and testing again. If it's as much moisture as I suspect, you might have to do this twice.
     
  8. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    Doug,

    I found the instructions on how to recharge the desiccant and will keep you inform on its progress.

    Again, thanks in advance.

    Sam
     
    Last edited: Oct 30, 2025
  9. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    8,969
    Location:
    Earth
  10. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    Colin,

    Great idea, just purchased one on line....I do hope this fixes the issues with the camera. Will keep all posted on its progress.

    Cheers,

    Sam
     
  11. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    I just replaced the desiccant and was wondering how long should I wait till I check the camera?

    Again, thanks in advance.
     
  12. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    8,969
    Location:
    Earth
    24 hours before cooling the camera. And maybe bake it and repeat.
     
  13. Sam Medina

    Sam Medina Standard User

    Joined:
    Nov 12, 2015
    Messages:
    7
    Gentlemen,

    Thank you for your assistance, but I had no luck. Recharged (baked) the desiccant four times with no luck at all. I believe my next step is to ship the camera of for servicing.

    Please advise if there is something else I might be able to do.

    Cheers,

    Sam
     
  14. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    8,969
    Location:
    Earth
    Nope, you'll need to request the RMA, in the link above.
     