Staff Assigned: Bill Lynch Goofy STi image all of a sudden

Discussion in 'STF Series CCD Cameras' started by Larry Groom, Oct 7, 2020.

  1. Larry Groom

    Larry Groom Cyanogen Customer

    Joined:
    Oct 15, 2014
    Messages:
    38
    My STi cameras started giving images as shown below. There's a weird "herring bone" pattern in the
    middle half of the image. I swapped USB cables, swapped computers, long exposure, short exposure, Bin 1, Bin 2 all the same and gave up. It almost looks like the anti-blooming feature isn't working in the distorted part of the image.

    Any suggestions? (I already emailed Bill)

    Larry
    lobo59@me.com

    Bad STi Image.jpg
     
  2. Larry Groom

    Larry Groom Cyanogen Customer

    Joined:
    Oct 15, 2014
    Messages:
    38
    I forgot to include a Dark frame with the same weird pattern.

    Larry
    lobo59@me.com
     

    Attached Files:

  3. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    7,429
    Location:
    Earth
    Hi Larry, can you post a 1x1 FITS light image, bias, and dark ? We can't do anything with a JPG.
    Also, what's the optical configuration - eg is it in an OAG-8300 or some other off-axis gizmo, or in a guide scope, etc.
    Did you visually inspect it for contamination on the optical window?
     
  4. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    9,967
    "Weird" is right!

    Will definitely want to see FITS frames for this one. Some thoughts based on the JPEG images:

    I can see a loss of charge transfer efficiency vertically in the distorted area, in addition to the horizontal lines - hot pixels are streaking vertically. That implies that the CCD clock voltages are unstable.

    My best guess is that an internal power supply regulator has failed. The voltage could be bouncing up and down as the vertical and horizontal clocks start and stop, resulting in out-of-spec clocks and producing that pattern.
     
  5. Larry Groom

    Larry Groom Cyanogen Customer

    Joined:
    Oct 15, 2014
    Messages:
    38
    Here are Light, Dark and Bias FITS frames at 3 second exposures for the Light and Dark taken with a cover on the camera, ie not installed on a telescope. I first noticed the issue when the camera was installed in a OAG-8300 attached to a STF-8300. The results were the same as these new frames. The frames taken with the STi installed on the camera are on the computer at my remote observatory. I'll try to download those later in case you need them.

    Also I confirmed that the drivers and firmware are up to date.

    Larry

    PS I was able to download the images taken with the camera on the scope and have added them to this post including one with a dark subtracted from the light.
     

    Attached Files:

    Last edited: Oct 7, 2020
  6. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    9,967
    Looks like your camera needs to go in for service.

    Please contact Bill @ SBIG Service & Repair, bill@sbig.com
     
  7. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    7,429
    Location:
    Earth
    Hi Larry, the camera definitely has an electronics failure as Doug outlines. I suggest you get this in to @Bill for repair.
     