Dear all from SBIG TEAM, Suddenly my SBIG ST10XME is showing in a permanent manner -152 degrees in SGP PRO.. is this related to any driver issue or to SGP ? Permanent manner means, at the moment the CCD is turned on and SGP is showing the value -152 hence, the system is not cooling anymore.. And could it help ? if I provide to you any logs captures via SBIG DRIVER CHECKER ?? r do you think the matter is related to SGP ? Any ideas are welcome.. few weeks ago, the correct values were shown in SGP.. SGP = sequence generator pro Thank you Martin
Thank you Doug, I will try to find a copy of CCDOPS.. is there any special process I need to stick to ? and one question, are the MaxImDLObsoleteDriver needed where the SBIG CCD is only operated by SGP PRO ? I have this new LAPTOP and would rather only use the SBIG CCD with SGP PRO.. Do you think the OBSOLETE DRIVERS are needed to be installed beside the standard drivers provided with SBIG DRIVER CHECKER 64 ?? for the combination above ? Thank you Martin
Did find CCDOPS Ver 3.11B4, Nov 2011 hope the version is still good enough to make the functional test..
No, use the current CCDOPS. That one is there to support the really old parallel port cameras. You can download the current version from the product page of any of the STF cameras. MaxIm DL Obsolete Driver pack is not required.
That temperature reading is usually the result of an open thermistor. The thermistor becoming open is usually the result of it being corroded to the extent that the leads become disconnected (eaten away). -Bill
Hello Bill, here is the result from the CCDOPS and indeed, also CCDOPS shows -153.65"C That was not the case last week.. when I did have the same ccd in action with MAXIM DL.. See the FIT HEADER from the DARK FILE taken last week.. it was still showing correct tempretaure values.. SIMPLE = T BITPIX = 16 /8 unsigned int, 16 & 32 int, -32 & -64 real NAXIS = 2 /number of axes NAXIS1 = 2184 /fastest changing axis NAXIS2 = 1472 /next to fastest changing axis BSCALE = 1.0000000000000000 /physical = BZERO + BSCALE*array_value BZERO = 32768.000000000000 /physical = BZERO + BSCALE*array_value DATE-OBS = '2018-10-15T17:10:08' /YYYY-MM-DDThh:mm:ss observation start, UT EXPTIME = 10.000000000000000 /Exposure time in seconds EXPOSURE = 10.000000000000000 /Exposure time in seconds SET-TEMP = -20.000000000000000 /CCD temperature setpoint in C CCD-TEMP = -12.705067188926705 /CCD temperature at start of exposure in C XPIXSZ = 6.7999999999999998 /Pixel Width in microns (after binning) YPIXSZ = 6.7999999999999998 /Pixel Height in microns (after binning) The ony difference between last week and today is.. I did only exchange the descicannt and that is it..and suddenly the CCD shows at the moment connected -153.65"C..and one more issue, I did also send the CCD to a friend " very well packed " since he wanted to use it and see how " GOOD " the CCD is.. " you know, many dream of having one SBIG ST10XME " Is this a case which needs to be repaired by you Bill ? Cheers Martin
Hello, thank you all (especially Martin!) for the fast and uncomplicated help. I´m the guy who is dreaming about having an ST10XME in the observatory After all tries with with different drivers and software, i'm also at the point, that this might be an hardware issue. Well, here at work, we have an electronics workshop. Repairs at this (and higher) level (Thermistor replacement) should be easily possible. But i would only try that check or replacmenet, when you (Bill) can give some instructions where the sensor is located and what type it is. Another or may be related question according temperature: The camera gives also an environmental temperature. This value did also not change over the 2 hours yesterday evening. Always 25 degree, but it should read something around 19 degree. Is this a real measured temperature? Do you have any advice on this? Cheers Torsten
The camera should come in for repair. Replacing the thermistor involves more than simply replacing a single component. The pickoff mirror and ccd have to be removed. Reassembly is a process of its own. Part of which involves small wire soldering, parfocal alignment, ccd cleaning and desiccant refresh. Send me an email, bill@sbig.com, if you're interested in sending it in. -Bill
Thank you Bill, I will contact you directly for the repair service.. I think I will do that.. the SBIG ST10XME is a very good ccd to leave it unused.. Still, very unusual, why one week earlier the CCD did work and then after sending it not " despite well packed ".. due to "open thermistor " could that be due to transportation ?? Cheers Martin
I can only assume that the thermistor connection hadn't broken loose yet, if in fact this is the cause of failure. The symptom, a temp display of -153.65, indicates that there is no temperature being sensed by the electronics. This reading, -153.65, is the default temp display with no temperature sensor installed. Therefore, the camera isn't sensing the thermistor. Historically, when I've encountered this scenario, the thermistor has been corroded to the point that one or both of the small wires soldered to it has been eaten away. If this indeed is the case, I wouldn't be surprised to see corrosion on other parts inside that section of the camera as well. Why it worked a week prior? I can't be certain, things work until they don't. I'm merely relaying what I have seen historically regarding this symptom. -Bill
Did send you email with questions..how to proceed to send the CCD to you for service..thank you Martin