STL11000 - Horizontal banding w/either guider chip?

Discussion in 'Legacy Models - Community Support' started by Ishtim, Aug 21, 2024.

  1. Ishtim

    Ishtim Standard User

    Joined:
    Jul 7, 2015
    Messages:
    7
    Hello all,

    Both Internal & External guide chips for my STL11000 began exhibiting horizontal banding in the images. When tested at night, I am unable to discern guide stars in either chip's images, just banding. LEDs do not indicate any errors, but I tried swapping out a the PSU with my older/working 6303 PSU and see no difference.

    Main imager functions flawlessly.

    Are there any recommended troubleshooting tips?

    Ive included FITS taken with CCDOps using each guider chip and both power supplies. These were taken in a dimly lit room without stars.

    Thanks in advance!
    Andrew
     

    Attached Files:

  2. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    7,962
    Location:
    Earth
    Are the bias and dark frames clean?
    Try powering everything off, and disconnect the Remote Guide Head (RGH) cable on the main camera body.
    Then test the internal guide sensor.
    It could be a bad RGH cable.
     
  3. Ishtim

    Ishtim Standard User

    Joined:
    Jul 7, 2015
    Messages:
    7
    Thanks Colin,
    I've completed the recommended steps and both bias and dark look OK to me (see attached FITS).
    *Neither bias nor dark frame were taken in complete darkness (observatory venting lets in some light during day), but the banding is not visible.
    The FITS attached previously above were all 1 sec. exposures, the FITS attached here were 0.12 and 3 sec for bias and dark respectively.
    I have another RGH and extra cable I will try out this evening and report back...
     

    Attached Files:

  4. Ishtim

    Ishtim Standard User

    Joined:
    Jul 7, 2015
    Messages:
    7
    Hmmmm, I've swapped RGH cables and re-shot bias and dark with both chips, banding is back.
    Images attached.

    Going to swap out RGH's now.
     

    Attached Files:

  5. Ishtim

    Ishtim Standard User

    Joined:
    Jul 7, 2015
    Messages:
    7
    Using a different RGH has solved the "mystery" banding. :) (see attached)

    Is the RGH a serviceable item or do I have a new paperweight?
     

    Attached Files:

  6. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    7,962
    Location:
    Earth
    Turn the cooling on and see what you get - it will increase the power supply load.
    We cannot service any of the STL, ST, or STT cameras - no parts exist.
     
  7. Colin Haig

    Colin Haig Staff Member

    Joined:
    Oct 27, 2014
    Messages:
    7,962
    Location:
    Earth
    Andrew, one other possible cause is corrosion in the chamber on the electrical contacts.
    This can be caused by moisture - the desiccant is no longer doing its job.
    Replacement desiccant for the STL is still available: https://diffractionlimited.com/product/replacement-desiccant-plugs/
    Disassembly and cleaning/removal of corrosion is not something we would be able to do, as parts arent available.
     
  8. Ishtim

    Ishtim Standard User

    Joined:
    Jul 7, 2015
    Messages:
    7
    Thanks Colin,
    I am in the humid southeast US and plan on doing some "surgery" in the near future.
    I bake desiccant plugs often...
    Andrew
     