STXL-11002 with FW8G-STXL - issue with flats and with reliably connecting to camera

Discussion in 'STX and STXL Series Cameras' started by Tiffany, Nov 14, 2023.

  1. Tiffany

    Tiffany Standard User

    Joined:
    Jan 19, 2023
    Messages:
    8
    I have an STXL-11002 CCD camera with an attached filter wheel FW8G-STXL that for the last ~year has been giving me intermittent issues, resulting in Camera Error (17) when trying to expose using MaxIm DL. Sometimes, this issue is resolved by turning the power off and back on again, but sometimes I lose entire clear nights because the camera has issues exposing. I thought it may have been a weird filter wheel issue - see previous thread - and it may still be, but I need some help.

    Just today I updated the flat field images for calibrating and there are white blobs and streaks across the edges of the flats, and there is a different pattern produced with each of the filters. Flats were taken today during the day in a closed dome with a Flat-Man flat panel. I did a whole bunch of flats today in two separate instances - the second set is worse than the first! Unprocessed images from overnight last night, however, look normal.

    What is happening with my flat field images, and could it be related to the intermittent issues I have taking exposures or likely two separate issues?

    I've attached a couple of single flat frames as an example. flat-GRN2-0007.fit was done in the first round of flats today, while flat-BLU2-0005.fit was done in the second round of flats. I've also attached a raw light frame of M31 from exposures done last night in the middle of the observing session where I don't notice the same issue and I'm waiting to see what things look like tonight...

    I've just had the main camera on my telescope stop working (an Apogee CG-16M) and have been relying on the SBIG camera much more than usual. (I am also in the market to buy a new CCD.)
     

    Attached Files:

  2. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    10,316
    Error 17 is "Shutter Error". So that's a mechanical or electrical issue with the camera's shutter. It may require service.

    The blobs are ice. You need to recharge the desiccant. Instructions are in the user manual.

    These days most cameras are CMOS. Please have a look at our web site!
     
  3. Tiffany

    Tiffany Standard User

    Joined:
    Jan 19, 2023
    Messages:
    8
    Thanks, Doug! Quick and straightforward thoughts on both issues, I appreciate it. I'll look into recharging the desiccant.

    For maybe requiring service for the shutter issue, I'd be interested in knowing more though wouldn't be able to do that until I had a back-up camera in place at the Observatory here, as long as it continued to sometimes-function, at least.
     
  4. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    10,316
    When you're ready, let us know and @Tim will provide an RMA and instructions. Hopefully it's something simple like a sensor misalignment.
     
  5. Tim

    Tim Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    1,869
    If you would like to send it in for service, please submit the following RMA contact form. Once I receive your submission, I can provide you with an RMA number and return instructions:

    http://diffractionlimited.com/request-rma
     

Share This Page