STF-8300M dessicant plug

Discussion in 'STF Series CCD Cameras' started by Moon Watcher, Nov 18, 2014.

  1. Moon Watcher

    Moon Watcher Cyanogen Customer

    Joined:
    Oct 10, 2014
    Messages:
    10
    I've only had my camera for less than a month and believe it started frosting up on me. I was checking the instructions, in the manual about how to recharge the desiccant. I unscrewed the knurled plug/bolt and looked at the plug and found the 0-ring. Under the plug was some something white that I figured was the desiccant. The question... there is mention, in the manual, of another plug to seal the chamber while heating the other plug. I could not find any other part. If I use this plug, then what do I use to cover the chamber? Am I missing something?

    SBIG STF-8300 Manual

    1. Unscrew the desiccant container from
    the side of the camera opposite the
    connectors and remove the O-ring.
    2. Plug the resulting hole by screwing in the supplied bolt or plug 2 or 3
    turns.
    Finger tight is adequate. Don't put a wrench on it.
    3. Heat the desiccant container in an ov
    en at 350°F (175 deg C) for 4 hours.
    The solder used to seal the can melts at
    460 degrees F, so be sure to stay at
    least 50 degrees below this number. Preheating the oven to avoid hot spots
    is advised.
    4. Replace the desiccant container into cam
    era, being careful to reinstall the
    O-ring and insure that it does not get pinched.
     
  2. Antoine Vergara

    Antoine Vergara Cyanogen Customer

    Joined:
    Oct 14, 2014
    Messages:
    38
    Location:
    Lentin, France
    Hi,
    In the camera operating manual it's indicated "
    Place a small piece of electrical tape over the hole
    in the side of the CCD chamber while you are baking the desiccant plug to keep unwanted dust and
    moisture out of the chamber. "
    Antoine
     
  3. Moon Watcher

    Moon Watcher Cyanogen Customer

    Joined:
    Oct 10, 2014
    Messages:
    10
    Hi Antoine,
    Thanks for the reply, but my manual does not mention "electrical tape". That's why I highlighted the section about "supplied bolt or plug".
    Putting a piece of tape over the hole should work, but I thought I was missing a part.
    John
     
  4. Antoine Vergara

    Antoine Vergara Cyanogen Customer

    Joined:
    Oct 14, 2014
    Messages:
    38
    Location:
    Lentin, France
    Hi John,
    In fact it's in the STT 8300 operating manual, not the STF.
    Sure it will work just fine.
    I generally wrap the dessicant in a cooking aluminium sheet, the time to go from the oven to the observatory.
    Antoine
     
  5. Moon Watcher

    Moon Watcher Cyanogen Customer

    Joined:
    Oct 10, 2014
    Messages:
    10
    Hi Antoine,
    Good thought about using the aluminum foil. Thanks again.
    John
     
  6. ctrleng2

    ctrleng2 Standard User

    Joined:
    Nov 30, 2015
    Messages:
    2
    I just bought an STT-8300 from a friend. I ended up baking the desiccant plug o-ring. Obviously I need to replace it. Does anyone have a McMaster Carr part number or a specific size and type for the replacement? Many thanks!
     
  7. Doug

    Doug Staff Member

    Joined:
    Sep 25, 2014
    Messages:
    9,968
    McMaster Carr 9557K463. $7.70 will get you a ring plus 99 spares...
     
  8. ctrleng2

    ctrleng2 Standard User

    Joined:
    Nov 30, 2015
    Messages:
    2
    Many thanks!
     